Discussion in 'All Categories' started by Mohammad atieh - Feb 27th, 2018 11:32 am. | |
Mohammad atieh
|
Dear sir If you can help me to know what is the proper way to use cidex solution as disinfectant for laparoscopic instrument regarding time and number of cycle that we should use Thanks |
re: Cidex solution
by Dr. R. K. Mishra -
Jun 12th, 2019
6:35 pm
#1
|
|
Dr. R. K. Mishra
|
Glutaraldehyde is normally used as a 1–2% solution for the cold disinfection of various types of endoscopes and probes used in laparoscopic surgery. Instruments should be left immersed in 5 litre glutaraldehyde solution at least for 6 hours. It should be used only 15 times or 15 days whatever comes earlier. If the solution become dirty and black even before it should be discarded. If the instruments are not properly cleaned and dirty blood stained instrument is emerged in cidex it will destroy the entire solution and then you have to change the cidex. |